Calibration of NI 9229/9239
I found this procedure (http://www.ni.com/pdf/manuals/372263d.pdf) to calibrate the NI 9229/9239 we have at work. Is it possible that, in the gain matching verification, step 3. b), when we are told to refer to table 3, it should be table 8? The same goes for step 4. b) in the phase matching verification.
Thank you!
Elyse Boulanger
(2,379 Views)
Hey Elsietheengineer
Could you give me a little insight into why you think that might be the case?
Austin Stanton
(2,334 Views)
Hi!
Tables 8 and 11 are located right below the line where a reference is made to refer to table 3 while there is no reference to refer to tables 8 or 11 anywhere in the text. Tables 8 and 11 have the same content, but are both different from table 3.
Thanks!
(2,318 Views)