Hi LargeT,
E-Series cards can count quadrature encoder signals, but are not capable of making use of the Z-index (wheras M-Series cards and TIO cards can). This distinction is most likely the source of the apparent conflicting information.
Here is a tutorial that describes the process of using an E-Series card with a quadrature encoder.
One other piece of information to note is that E-Series cards cannot make use of the paired differential outputs of your encoder. You can either simply not connect the ~ signals, or you can use external circuitry.
Hopefully this will clarify the situation. Let me know if you have more questions as you move forward.
Have a great day!
Travis W
01-25-2007 09:55 AM
01-28-2007 11:34 PM - edited 01-28-2007 11:34 PM
Message Edited by Travis W on 01-28-2007 11:34 PM
01-29-2007
09:58 AM
- last edited on
07-11-2025
10:32 AM
by
Content Cleaner
Hi LargeT-
In case you have access to LabVIEW 7.1 or later you should consider using NI-DAQmx.
Hopefully this helps-
02-13-2007 09:11 AM
Thank you all for you input.
We finally got the system going based on the information and examples you provided. The final solution lay in the Up/Down Counter VI in the discussion you referanced, and a little re-wiring on our part.
Thanks again for everything!!!
LargeT