Cannot measure AC current using NI VirtualBench - NI VB-8012

Solved!
Go to solution

Hi All,

 

I’m trying to perform manual AC current measurements, but a result is just a noise.

 

I have a simple series circuit with FGEN, as VirtualBench, and load resistor of 50[Ohm]. My measuring setup is as described in the attached image. Except the batteries – I use fGEN at frequency of 486kHZ.

I’m pretty sure that my measuring setup is correct, because when I switch to DC mode – I have a comprehensive current measurements. In addition, when running in AC mode, through the AC multi meter connection, the load resistor is warming up – which means that there is a current in my circuit.

Both fuses were tested for continuity – and were fine.

 

What do I miss here? Is there some kind of “reset” for the virtual bench?

 

Thanks in advance

Art

0 Kudos
Message 1 of 3
(3,203 Views)
Solution
Accepted by topic author ArtB83

ArtB83 wrote:

 

I’m trying to perform manual AC current measurements, but a result is just a noise.

 

I have a simple series circuit with FGEN, as VirtualBench, and load resistor of 50[Ohm]. My measuring setup is as described in the attached image. Except the batteries – I use fGEN at frequency of 486 kHz.


I'm not an analog engineer, but I hypothesize that this circuit's frequency is beyond what the VirtualBench DMM can measure.

 

Looking that the specs, it appears that the DMM has guaranteed accuracy for AC signals up to 5 kHz:

vbdmm.png

 

If you reduce your circuit's frequency to 5 kHz or lower, do you see good measurements?

Joe Friedchicken
NI Configuration Based Software
Get with your fellow OS users
[ Linux ] [ macOS ]
Principal Software Engineer :: Configuration Based Software
Senior Software Engineer :: Multifunction Instruments Applications Group (until May 2018)
Software Engineer :: Measurements RLP Group (until Mar 2014)
Applications Engineer :: High Speed Product Group (until Sep 2008)
Message 2 of 3
(3,177 Views)

Hi Joe, 

You were 100% right - When I dropped my work frequency to 5kHz - the ampere-meter started to show correct values. 

Thank you very much.

 

Art