Ch-Earth Ground Isolation

   Dear Sir/Madam,
 
What is the difference between the "Channel Earth Isolation" (found mostly in newer USB DAQ products) and the other 2 types of isolation (ch-ch and bank)?
 
Thanks!
 
Regards
Albert
0 Kudos
Message 1 of 4
(4,078 Views)

Hello Albert,

-  Channel Earth Isolation is the isolation between a board’s channel and the earth ground of the board.  Isolation is a rating of the maximum voltage differential that can occur between these two.  This type of isolation is also a part of bank and channel-to-channel isolation.

-  Bank Isolation is where groups of channels form three “banks” (analog I/O, digital input, and digital output) that have a separate ground plane from each other and earth ground. 

-  Channel-to-Channel Isolation is isolation in-between analog I/O channels and bank isolation for digital I/O channels. 

I found the screenshots in the product manuals located on the product page under the Resources tab.



Message Edited by samantham on 11-28-2007 03:23 PM
Samantha
National Instruments
Applications Engineer
Download All
0 Kudos
Message 2 of 4
(4,045 Views)
   Dear Samantha,
 
Thank you for the explanation.
 
Let me just re-confirm. The channel-earth isolation means that all the banks are not isolated from one another, but all of them (as a group) are isolated from the earth (PC earth). Is this correct?
 
Then there is another question - is the analogue ground isolated from the PC earth (say, I want to use differential inputs)?
 
Thank you.
 
Regards
Albert
 
0 Kudos
Message 3 of 4
(4,033 Views)
 

Hello,

 

 

You are correct about the channel-earth isolation.  Yes, the AI GND / AO GND channels are isolated from the PC ground.  In the first Figure form the previous post, you can see from note 2 that Vc includes AI GND.  In addition, there is another resource for a USB isolation diagram.  The link is provided below and the diagram is Figure 11-1.

 

NI USB-621x User Manual    

Samantha
National Instruments
Applications Engineer