Ch-Earth Ground Isolation
   Dear Sir/Madam,
What is the difference between the "Channel Earth Isolation" (found mostly in newer USB DAQ products) and the other 2 types of isolation (ch-ch and bank)?
Thanks!
Regards
Albert
(4,173 Views)
Hello Albert,
-  Channel Earth Isolation is the isolation between a board’s channel and the earth ground of the board.  Isolation is a rating of the maximum voltage differential that can occur between these two.  This type of isolation is also a part of bank and channel-to-channel isolation.
-  Bank Isolation is where groups of channels form three “banks” (analog I/O, digital input, and digital output) that have a separate ground plane from each other and earth ground.
-  Channel-to-Channel Isolation is isolation in-between analog I/O channels and bank isolation for digital I/O channels.
I found the screenshots in the product manuals located on the product page under the Resources tab.
Message Edited by samantham on 11-28-2007 03:23 PM
Samantha
(4,140 Views)
   Dear Samantha,
Thank you for the explanation.
Let me just re-confirm. The channel-earth isolation means that all the banks are not isolated from one another, but all of them (as a group) are isolated from the earth (PC earth). Is this correct?
Then there is another question - is the analogue ground isolated from the PC earth (say, I want to use differential inputs)?
Thank you.
Regards
Albert
(4,128 Views)
You are correct about the channel-earth isolation.  Yes, the AI GND / AO GND channels are isolated from the PC ground.  In the first Figure form the previous post, you can see from note 2 that Vc includes AI GND.  In addition, there is another resource for a USB isolation diagram.  The link is provided below and the diagram is Figure 11-1.
NI USB-621x User Manual
Samantha
(4,112 Views)