Hey Fair Hope,
You are correct that when using CJC, AI0 and AI8 are not available for use. You can read some more information about this below.
How Do I Use the Cold Junction Compensation on the SCB-68?
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019MnvSAE&l=en-US
Connecting the thermocouple as differential should show some improvements in the noise. What DAQ device are you using? I am assuming you are connecting the thermocouples to channels 14 and 15 in RSE. It seems like your only real options are to either remove some of the connections that are in use or upgrade your device to more channels. You could perform a test by disconnecting some of the other connections in use and connect the thermocouple as differential to see if it does in fact improve the noise.
I have also linked a couple of articles that discuss improving noise.
Reducing Thermocouple Noise to Improve Measurements
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P8wxSAC&l=en-US
Field Wiring and Noise Considerations for Analog Signals
https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...