Channels got influenced by each other

Hi guys,

 

I'm quite new to LabVIEW, recently i'm setting up a small system to measure the pressure and temperature simultaneously. A PCI-6071E card is used, AI0 and AI1 are connected to pressure transducer and thermocouple output respectively. The actual voltages of the two channels should be 1.67 and 0.23 . When I activate only one input, both signals are accurate. But when I activate the two signals together, the pressure terminal reading is OK as 1.67, but the temperature terminal reading will increase to 0.26(but the actual voltage is still 0.23 as measured by multimeter), when i deactivate the pressure transducer, the temperature voltage will be 0.23 again.

 

Anyone can tell me why?

 

Cheers,

Pengcheng

0 Kudos
Message 1 of 4
(4,425 Views)

Sounds like the MUX is not given enough time to settle before the ADC takes it measurement.  Try slowing down your sample rate.  Another trick would be to add a channel that is grounded and have the DAQ measure that grounded signal between your two channels.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 4
(4,419 Views)

Thanks for your reply. I set the sample rate to 1Hz and also tried the other way, but I'm afraid that they didn't work. It's reaaaly weird.

0 Kudos
Message 3 of 4
(4,412 Views)

Hi,

 

The issue sounds like ghosting between channels. Please find the following resources which may solve this issue:

How Do I Eliminate Ghosting from My Measurements?

 

Troubleshooting Unexpected Voltages, Floating, or Crosstalk on Analog Input Channels

 