Common mode voltage in NI 6250
HI,
What is the limit for the common mode voltage when using differential inputs in NI6250?
IN my application the differential voltage is  few volts but with reference to ground, it is 80 volts.
Can it work?
Is this issue addressed anywhere?
Rafi
(2,993 Views)
Accepted by topic author Rafi2003
Hi Rafi,
The 6250 is not able to handle such a large differential to ground. I've found the specification in the NI 625x Specifications.
Maximum working voltage for analog inputs
(signal + common mode) ................ ±11 V of AI GND
If you need to have that high of a common mode rejection you might considering looking at some of our CompactDAQ modules. For example the NI9205 has 250V continuous Channel to earth ground Isolation.
Steven K.
Software Engineer
(2,973 Views)
