DC Voltage Measurement using C-Series
What is the recommended compact DAQ for measuring DC signals up to +/- 10V ?
I have two NI 9233 modules that fit into NI USB 9162 carriers, however these have hard-wired AC coupling.
There are thermocouple modules like the 9211A but they can only handle +/- 80mV.
Thanks.
(3,613 Views)
The best way to evaluate which cDAQ module is best for your needs would be to use our online advisor which can be found here.
Follow the steps of the advisor to view different module specifications to establish which module can perform the measurements you require at the best price.
Good Luck,
Tom
NIUK
(3,592 Views)
Thanks, I had looked in here.
It looks as if the 9205 is suitable but I can't see confirmation of DC coupling.
Could you please confirm for me?
(3,584 Views)
If you look at the manual for the NI-9205 here, and navigate to page 19 you will see that the input coupling is listed as DC:
Analog Input Characteristics
ADC resolution................................. 16 bits
DNL..................................................No missing codes guaranteed
INL....................................................Refer to the
AI Absolute Accuracy Tables and Formulas Section
Conversion time
       R Series Expansion Chassis ....... 5.2
μs or 192.3 kS/s
       All other chassis .........................4
μs or 250 kS/s
Input coupling...................................DC
Nominal input ranges........................±10 V, ±5 V, ±1 V, ±0.2 V"
Hope this helps!
Tom
NIUK
(3,577 Views)
Many thanks.
It's all there if you know where to look. 😄
(3,558 Views)