Hi NF1,
From my understanding you are currently using NI-4551 and NI-DSA in your application and would like to know what will change if you move to NI-4461. You are correct that the NI-4551 has reached its end of life and the recommended replacement is the NI-4461. The hardware driver that supports NI-4461 is DAQmx. If you upgraded your application to use the NI-4461 you would have to use the DAQmx function calls to write your application. This would be the most notable change.
You can download the DAQmx driver for free online at this link. The driver will include example programs written in C++. See the DAQmx Readme for the path to the examples. If you have XP, you can find the DAQmx read me by going to Start>>Programs>>National Instruments>>NI-DAQ>>NI-DAQ Readme.
I hope this helps!
01-18-2008 08:43 AM
Hi Ima,
I've taken a look at the DAQmx API and the functionality doesn't seem to include the same ability as the DSA API -- e.g. functions to run a Swept Sine wave analysis and getting magnitude and phase plot responses. The DAQmx drivers just have basic functions to configure, read, write, etc to the NI4461.
For the NI4551 card, I use the legacy/traditional DAQ C++ drivers and the DSA C++ drivers. The DAQmx driver just replaces the legacy/traditional DAQ drivers, but I have not found an equivalent DSA driver upgrade for the NI4461.
Thanks,
Nia
01-22-2008 10:24 AM
01-22-2008
11:49 AM
- last edited on
04-29-2024
02:10 PM
by
Content Cleaner
Hi Nia,
You are correct that DAQmx replaces traditional NI-DAQ. NI-DSA and the old DSA boards are different from NI-DAQmx and the new DSA boards because computers became fast enough to do signal processing in software. For this reason, the newer DSA boards no longer do on board signal processing. The new boards focus on acquiring data and the signal processing is now done in software. The Sound and Vibration toolkit(SVT) was created to do signal processing in software. This decision was made to benefit customers because it is much easier to upgrade software than hardware.
That being said, for you to do complex signal processing, you will need Sound and vibration toolkit or measurement suite. The toolkit is made for LabVIEW. The measurement suite comes with a Sound and Vibration Assistant that is a stand alone, interactive program that allows you to do signal processing.
In regards to your question, can a LabVIEW DLL containing SVT VI’s be called from C++ project, I do believe this is a possibility. You would probably need to purchase the SVT and you can follow this Knowledge Base article on how to call a LabVIEW DLL in your C++ project.
I hope this helps!
01-23-2008 04:48 PM
01-24-2008
08:18 AM
- last edited on
04-29-2024
02:11 PM
by
Content Cleaner
We have DLLs for many of the audio measurement functions for C++ as well as systems engineering expertise to help you in the transition. We have multi-tone, THD+N and other DLLS. It is possible to build a DLL out of the SVT functions and export the measurements that you need.Send me an overview of your application and the functionality that you are looking for and I would be happy to help.
01-24-2008 02:06 PM
Hi Kurt,
The tester runs a Swept Sine Test and calculates the natural frequency and resonant rise of the DUT using the magnitude and phase data from the SweptFreqAxis MAGVIEW and UNWRPHASEVIEW. A list of custom test frequencies is used to run the test. Each station is able to run a number of different products that require different custom test frequencies, amplitudes, and number of test frequencies. When a different product is selected, the card is automatically reconfigured by the program.
As the swept sine test is run, the program polls the card to retrieve the Magnitude data so that the Magnitude plot can be drawn real-time. If real-time data retrieval is not possible, that's ok.
The test also sends a DC voltage to the DUT to determine its DC gain. However, I think this part of the test can be done using the existing DAQmx library.
These DSA functions are called :
NIDSA_init(resourceName, IDQuery, resetFlag, &DSAsession);
NIDSA_error_message(DSAsession, rc, errorMessage);
NIDSA_configure_dsa_mode(DSAsession, mode=SWPSINE_MODE);
NIDSA_configure_swept_sine_source(DSAsession, amplitude, autoLevelEnable=0, autoLevelChannel=1, idealRef=1, refLowerDeviationdB=-3, refUpperDeviationdB=3, maxLevel=10, rampingEnable=0, rampingRate=1);
NIDSA_configure_swept_sine(DSAsession, startFrequency, stopFrequency, numberOfSteps, repeat=0, linLog=1, sweepMode=2, autoMaxStep=512, fastThresholddB=1, slowThresholddB=6);
NIDSA_configure_swept_sine_custom(DSAsession, numberOfSteps, pTestFrequencies);
NIDSA_get_swept_sine_status(DSAsession, &sweepStatus, &measurementError, &overload);
NIDSA_set_input_voltage_range(DSAsession, channel, range=VRANGE_10V);
NIDSA_set_input_coupling(DSAsession, channel, coupling=COUP_DC);
NIDSA_configure_source(DSAsession, sourceOn, type);
NIDSA_configure_sine_source(DSAsession, freq1, amp1, freq2, amp2, dcOffset);
NIDSA_check_new_measurement(DSAsession, newMeasurementFlag);
NIDSA_get_measurement_length(DSAsession, type, 0, pNumMeasurements, &realOrComplex);
NIDSA_read_measurement(DSAsession, type, startIndex, numMeasurements-startIndex, view, dbState=DBON, pkrmsUnits=PEAKUNIT, phaseUnits=DEGREES, &x0, &dx, fMeasurements);
NIDSA_close(DSAsession);
The hardware setup is pretty straightforward :
Analog Ouput Ch 0 --> Amplifier IN / DUT input
Analog Input Ch 0 --> Current Probe Output / DUT current input
Analog Input Ch 1 --> Capacitance Probe Output / DUT response
Ideally, as many of the configuration and test parameters as possible would be input parameters to the dll functions so that if the config or test needs to be changed the dll doesn't need to be modified or there don't need to be multiple dll's running. At minimum, the number of test frequencies, array of custom test frequencies, and amplitude would need to be input parameters.
Thanks,
Nia
01-24-2008
03:39 PM
- last edited on
04-29-2024
02:11 PM
by
Content Cleaner
01-31-2008 11:16 AM - edited 01-31-2008 11:16 AM