I'm looking at two Ethernet cdaq chassis connected together, one 4 slot cDAQ-9185 and one 8 slot cDAQ-9189. Was looking at populating them with NI-9234. My quick less educated crude calcs show the bandwidth should be enough to capture 48x channels (12 slots with 4 channels each module) @ 24 bit and 51.2kHz? Although I've likely made some mistakes.

0 Kudos
Message 1 of 4
(1,769 Views)
Solution
Accepted by topic author PeterBrown

You're in the right ballpark for a raw calculation.  I'm assuming the 24-bit samples will be carried in a 32-bit integer though, which brings the bandwidth demand up to almost 80 Mbps.  If you're on a 100 Mbps connection, that may be too high a %age of the theoretical max to count on.  Others can probably comment with more complete knowledge.

 

 

-Kevin P

ALERT! LabVIEW's subscription-only policy came to an end (finally!). Unfortunately, pricing favors the captured and committed over new adopters -- so tread carefully.
0 Kudos
Message 2 of 4
(1,753 Views)
Solution
Accepted by topic author PeterBrown

Related article to calculate bandwidth - https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019QYVSA2&l=en-US

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
Message 3 of 4
(1,746 Views)

Thanks. That example in the link was for the same model of module so good to validate my rough calc

Which yes I only used 24 bit.. So explains why it was even more comfortable. I'll mostly be using them plugged into a gig ethernet port.