I am using PCI 6040E, LabVIEW 8.6, BNC 2090 for the data acquisition. I want to analyze EMG signals for developing a prosthetic hand. I am new to this thing. Can any one please suggest:
1. Is my DAQ system alright to do these measurements?
2. What additional hardware do I require (Besides EMG surface electrodes, of course for example do I require some amplifiers etc?
Many thanks.
06-25-2009 03:21 PM
EMG can range up to about 10kHz, so your acquisition system has that covered.
I'm pretty sure you will need some instrumentation amplifiers to scale up the signal at least a factor of 10. More importantly, it will give you some isolation between the electrodes and the acquisition system.
06-26-2009 10:49 AM
You definitely need a medical-grade amplifier appropriate for acquiring EMG before the DAQ card. You need both gain, safety (usually isolation, but at least low-leakage and protected inputs), an AC-coupled input (blocks DC offsets from electrodes and movement artifact), and filtering (to prevent aliasing of high-frequency noise).
You can find more information about this at our Biomedical User Group here: http://decibel.ni.com/content/groups/biomedical-user-group
Look in the Wiki or Documents area - I posted a note about sources for amplifiers like these.