Error Code: -10842 with calibration executive

We purchased Cal Executive 2.0 with the hardware/software for calibrating E-series multi-function DAQ cards a few months ago, and I finally tried it out.

1) I have a HP 53131A Counter, HP34401A DMM, and Fluke 5700A calibrator. Each time I run the Cal Exec, I must manually select my cal hardware ((a)does not find the 5700 and no option to select hardware - doesn't usi it if I select the address, (b) does not find the DMM but works if I manually select type and address, (c)does not find the 53131A counter and no option to select hardware - doesn't us it if I select the address)
2) Cal a PCI-6071E: I have never been past the section of 'Performing internal calibration'. I ALWAYS get the following:
"An UNKNOWN interface e
rror occurred at:
Step: 'Internal Calibration' in
Sequence: 'Verify E Series AI'.

Specific error information:
Error Code: -10842
Error Message: UNKNOWN ERROR"

3) Cal a DAQPad-6020E (USB): I have never been past the section of 'Performing internal calibration'. I ALWAYS get the same message as step 2).

After steps 2) and 3), I click terminate, and Cal exe is locked up, and must be terminated with task manager.

All cal equipment is connected via GPIB and responds to the IDN query via MAX. The PCI-6071E passes the E-series diagnostics web page of NI's web site (even the internal cal); the DAQPad-6020E does not get past the counter test since the application locks-up.

Machine specifics:
Compaq 7998 (1 GHz Athlon), Win2K, 256 M RAM, Cal Exec 2.0, VISA 2.6, IVI 1.83, DAQ 6.9.2, Teststand engine 1.03.100, CVI RTE 5.510.4.0 LV RTE 6.0.

Any Ideas? Money is very tight here, and at this point I am quite upset with the results thus far.

Todd Cazallis
Senior Function
al Test Engineer, Kimball Electronics-Auburn Indiana
0 Kudos
Message 1 of 3
(2,739 Views)
Hello Todd,

I am not sure if you have already gone through this link. If not, I would suggest you to take a look at it and please let me know if it helps.

Keywords Search: 10842


Please do let me know how it goes.

Sincerely,
Sastry V
Applications Engineer
National Instruments

Message Edited by Molly K on 02-19-2005 12:04 AM

0 Kudos
Message 2 of 3
(2,740 Views)
Yes, I have looked at all of the links.

I tried again a few days ago after installing another PCI-6071E into the PC, and it passes the internal cal point, but now, the original PCI-6071E I was having problem with passes the internal cal point also.

I still have the same problems as before with 1) the DAQPad-6020E USB (I have 2, and neither pass the internal cal point; they will not finish the E-series diagnostic utility on ni's web site either since it locks up when counter 0 is tested, which is before the internal cal is performed), and 2) the cal executive does not recognize the calibrater on the GPIB bus, 3) the cal executive does not recognize the 53131A counter on the GPIB bus.

I have another E-series PCI card I can try, but I still need th
e rest of the issues to be resolved.