Error when calibrating cRIO-9211
Cal Exec 3.2.1 is giving me problems when trying to cal cRIO-9211 units. This is the error message:
Error -200081 occurred at DAQmx Start Task.vi at step Verify Analog Input
Possible reason(s):
Sample rate exceeds the maximum sample rate for the number of channels specified.
Reduce the sample rate or the number of channels. Increasing the convert rate or reducing the sample delay might also alleviate the problem, if you set either of them.
Number of Channels: 2
Sample Rate: 7.140001
Maximum Sample Rate: 6.000000
Need to know what can be done to fix this, or need a patch ASAP.  Customers waiting on modules.
(3,281 Views)
Hey DTI1,
Thanks for contacting National Instruments support.  Do you have the required equipment for calibration?  The calibration procedure states that you need to have a 9172 cDAQ chassis for calibration to work.  If you do not have this chassis then you may get this error or other issues.
Please refer to the help for Calibration Executive under Device Calibration Procedures.  If you have further questions, let me know.
Kenn North
Principal Product Manager - Search, Digital Analytics
http://ni.com/search
(3,267 Views)