02-18-2009 10:14 AM
Hello,
I'm using a CompactDAQ Chassis with 2 NI 9237 modules for strain gauge measurements.
The maximum excitation voltage of my gauges is limited to 1V...
The 9237 supports internal voltages from 2,5 to 10V.
I want to connect an external voltage supply. Now I found the restrictions of 2V-10V in the datasheet for external devices. Is there any way of getting my 1V voltage into this system?
Karsten
02-19-2009 06:46 AM
Hi Karsten,
When external excitation is enabled on the NI-9237, the excitation ±terminals (pins 6 and 7) disconnect from the internal cDAQ backplanesource and connect to the 4 terminal EX ± AI Bridge on the front of thedevice. With external excitation enabled, a 2-10 volt external powersupply needs to be connected in order to provide excitation.
02-19-2009 10:46 AM
I also tried to work with 2V external source... but then I get about 5 V between Pin 6 and 7? How could this happen?
I connected the external source to 2 of the 4 ex-Pins
I switched the settings in the vi to external source and excitation value = 2
02-20-2009 06:38 AM
Karsten,
this happens because of the limiation of the ADC in the NI-9237 which requires at least 2V.
regards
Tobias Daub
02-20-2009 06:42 AM
OK ...
up to 2V - I understand... but
why does it happen with my external source with 2V? Or with 2.1 V or 2.2 V?
The described effect disappears after an external voltage of 2.5V