Getting voltage spikes from an analyzer on the usb-6212 analog input

I have an NI USB6212 the BNC style DAQ and am trying to read in the voltage from a gas analyzer. I keep seeing these voltage spikes in the data stream though, and they are not present when I plug the voltage out from the analyzer to just an oscilliscope. I attached some screenshots of what I am seeing. All this vi has in it, is a daq assistant set to read voltage at 100 samples per second for various rates, 10Hz, 100Hz, 1kHz. you can best see the spikes on the 1kHz or 10hz graph.  I have the analyzer connected through a UPS to try and get rid of some line noise but that shouldnt be a problem anyway if it did not show up on the scope.

 

I should also mention that this unit was just evaluated by the manufacturer and they saw some spikes but none as high as this so it must be a problem at the labview/daq end.

Download All
0 Kudos
Message 1 of 5
(4,706 Views)

These spikes could be indicating a EMC problem , and this could be due to ground loops and switched power supplies.  For high frequency spikes ground loops don't have to have a wired connection.. a capacitive coupling in the USB or PC supply is all they need 😕

 

Usually the scope input shield it directly connected to (protection) ground. While the USB DAQ AGND is more floating ....

 

So some test:  hook up the analyser on the scope: (you noted it looks fine)  

hook up on the scope AND the  DAQ , spikes on scope and/or DAQ?

 

Can you try a notebook with the USB-DAQ ?    With and without power cord? 

 

here  I posted a way that sometimes helps to get rid of the spikes...

 

Last exit: Use a median filter ....

 

 

 

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 5
(4,668 Views)

It is already running on a laptop and i have tried it plugged and unplugged. and with both the analyzer and the laptop plugged in to the battery supply. I did use a median filter but that only gets the positive spikes removed. the negative dips remain, how do i tweak that filter to eliminate them?

 

By scope to daq you mean via a bnc splitter?

 

My real concern with using filtering however is that I cannot trust the analyzer completely now. I have other manufacturer's analyzers here that do not have this issue and do not require filtering.

0 Kudos
Message 3 of 5
(4,662 Views)

@labview12110 wrote:

It is already running on a laptop and i have tried it plugged and unplugged. and with both the analyzer and the laptop plugged in to the battery supply. I did use a median filter but that only gets the positive spikes removed. the negative dips remain, how do i tweak that filter to eliminate them?

Filter length..

 

By scope to daq you mean via a bnc splitter?

 Yes 🙂

 

My real concern with using filtering however is that I cannot trust the analyzer completely now. I have other manufacturer's analyzers here that do not have this issue and do not require filtering.

 

Output impedance noted in the spec?


 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 4 of 5
(4,654 Views)

I have found that the USB DAQ devices are noisier than their non-USB counterparts.  To verify the source of the noise, you might put a load on a second analog input (1k ohm resistor) and see if the same noise is present.

 

I have tried a few different things to minimize this noise.  First, I did as you did and unplugged the power on the laptop, and this helped for one project.  For another project where I was doing long term acquisitions and couldn't unplug the laptop, I used a powered USB hub in the USB chain and supplied the DC power to the hub with a benchtop power supply.  I also bought an isolated USB hub from Sealevel.  I think internally it optically isolates the USB signals from the computer to all of the peripherals.

 

A median filter should have eliminated both the negative and positive spikes.

Randall Pursley