How do I connect a thermocouple to a BNC 2110?
I am using a BNC 2110 board connected to a PCI 6251 and I need to connect it to a type T thermocouple, but I am unable to get a sensible reading when I run the Measurement and Automation Explorer.  I want to set up a voltage differential channel so I can just read in the voltage difference on the thermocouple.
My first issue is connecting the thermocouple to the BNC board.  When I am trying to set up the channel in the Explorer and I look at the connection diagram, it says: "the connection diagram tool does not support the accessory on connector 0 of your device: BNC 2110." So I have to choose a Supported Acessory, usually I select SCB-68 (but the other option give the same result).  The problem with this is that this diagram says I need to connect to ports 34 and 68, which correlate to the same Analog input on the BNC 2110.  All the other channelsI can select have similar issues: either the ports are the same input on the BNC or one of the ports doesn't have a correlation on the BNC 2110.  So I'm not really sure where to attach the thermocouple wires...  I have tried multiple things, including just putting one of the wires in the BNC inpu that correlates to the ports and just putting the other wire in the next input; this causes the readings on the explorer to fluctuate very rapidly in the 0-1 mV range even when I heat the thermocouple junction, so I do not think it is the proper connection.
Help please... Thanks!
(4,440 Views)
Each analog input channel has two connections (+ and -) and correspond to the center conductor and shield of the BNC. Have you looked at the manual?
(4,431 Views)
Thanks!  That was the key I was missing.  Do you know what section of the manual this is addressed in?  I still don't really see it.  The thermocouples are definitely picking up changes in temperature now, although the readings in the MA Explorer are very small and unstable.
If you have any tips you can offer on improving the stabiliity or increasing the voltage that woud be great.
Thanks again for the help, much appreciated.
(4,421 Views)