How to advance the sampling rate of PCI-MIO-16E-4 in single-ended sampling mode?
Hi
I am using a PCI-MIO-16E-4 DAQ card to sampling pressure signals. It works in the mode of single-ended sampling. The program is developed on the platform of Delphi 6.0. I want to know the factors that are influencing sampling rate,and how to advance the sampling rate effectively. Thanks for your answers!
(2,464 Views)
The sampling rate for your PCI-MIO-16E-4 board is 500kHz if you are sampling only 1 channel and 250kHz divided by the number of channels if you are sampling on multiple channels. The main factors that limit these sampling rates are the resistances and capacitance at the beginning of the amplifier circuitry. I have also seen large source impedances reduce these sampling rates. Usually, if your source impedance is less than 10k ohms you should be fine. If you are a magnitude larger than that you might need to use a unity buffer to transmit the exact same voltage but acting as a much smaller source impedance. Here are some excellent resources:
Is Your Data Inaccurate Because of Instrumentation Amplifier Settling Time?
http://zone.ni.com/devzone/conce
ptd.nsf/webmain/B9B5D37A04C772B58625686500695C20?opendocument
How Do I Create a Buffer to Decrease the Source Impedance of My Analog Input Signal?
http://digital.ni.com/public.nsf/websearch/CF83426BC3AC514A86256C10005A4771?OpenDocument
Hope that helps.
Ron
(2,464 Views)