I am experiencing noise on my PCI-6238 card on all analog inputs.
My company is manufacturing a machine using LabView and a PCI-6238 multifunction board to handle all the I/O.  The PC is a Lenovo desktop.  We connected our analog inputs from our sensors.  The 6238 was reading the signals except that the signal was noisy.  We checked the signal with an milliAmp meter.  The signal is fine.  It is not changing as much as the Labview is indicating.  On the Labview NI Max we are seeing the signal as varying about +/-0.3mA on the graph.  The noise seems to be around 2-4hz.
I have tried the following with no success:
1)     Disconnected all wiring from connection block except for a Fluke loop calibrator that is outputing 4-20mA.  The signal is still noisy.
2)     Reseated the 6238 card.  The signal is still noisy.
3)     Removed and readded the 6238 card from the configuration.  The signal is still noisy.
4)     Connected the DC - on the loop calibrator to the analog input ground.  The signal is still noisy.
5)     Removed the 480 VAC from the system and are only operating the PC on a extension cord.  The signal is still noisy.
Any help with this would be greatly appreciated.
(3,519 Views)
There's a duplicate post that we will continue here.
Mark E.
Precision DC Product Support Engineer
(3,500 Views)
The PC, card, and machine are new.  I disconnected all external wiring and used a 4-20mA loop calibrator.  The signal is still the same amount of noise.  I then wired an analog output to one of the analog inputs.  The noise was the same or worse.  I was experiencing a 0.6mA amplitude on the noise with a constant output.  Today I can get a reasonable signal, but then it starts ramping and then rails.  Sometimes the signal is an occilation that spans from 4mA to 20mA.  My signals incoming are constant.
I even took the PC to another location(my house) and used the analog output to power the analog input.  The results were the same noisy signal.  I was trying to eliminate the fact that my office may have "dirty" power with harmonics on the lines.
I have the negative on the analog input tied to DC common and the AI GND tied to DC common.  I am concerned that there is a problem with the PCI-6238 card.  It passes all the tests and has been calibrated.
I have isolated the breakout board from all field wiring and the noise is still eradict.  Please offer any assistance.
(3,490 Views)
I meant to continue this duplicate thread at the other location where I posted. Let us please continue at the following link.
Mark E.
(3,469 Views)