Is it possible to use a 9217 to measure 30 kΩ thermistors?
I have a 9217 (100 Ω RTD, 24-Bit Analog Input Module).  Can it be used for measuring thermistors that have a resistance of 30 kΩ at 0C?  They require an excitation current of between 10-15 uA.
The 9217 module: https://www.ni.com/en-us/shop/model/ni-9217.html
(3,389 Views)
Without digging into the spec in detail, you have a current source of 1mA and you can measure a voltage in the range of 0-400mV...
Depending on the uncertainty you need the most simple solution is a good (stable, low tempco) resistor parallel to your thermistor.
Depending on your temperature range (thermistor value) and the max current you can choose a value for R to stay in the 0-400 Ohm range.
Greetings from Germany
Henrik
LV since v3.1
(3,377 Views)
It's not clear from the spec, but can the current source be changed to values other than 1 mA (such as 10 uA)?
While I have the hardware, I ordered the wrong USB sleeve, so I can't test this out myself until the new sleeve comes in.  For some reason, the 9217 is not available to be simulated in MAX, or I would just set up a simulated instrument and see what options are available in the task setup.
(3,369 Views)
Accepted by topic author JoshB
Hi JoshB,
The excitation current on the NI 9217 is fixed, not programmable.
To create a simulated NI 9217 in DAQmx, you need to create a simulated CompactDAQ chassis first. Then right-click on the chassis and there will be an option to configure simulated modules.
Brad
Brad Keryan
NI R&D
(3,363 Views)
