02-12-2008 10:38 AM
02-13-2008 03:00 PM
Hi Raghunathan,
On NI DAQ boards, the AIGND, AOGND, and DGND are all tied together and all referenced to you computer ground. Information about this can be found in the following Knowledge Bases:
Are the Grounds on My M
Series and E Series DAQ Devices All Tied Together?
http://digital.ni.com/public.nsf/allkb/02A451C0A88A3EAE8625702F0059C98E?OpenDocument
How Are E Series DAQ devices
Grounded?
http://digital.ni.com/public.nsf/allkb/10AF3F86965DA0CEC1256BBB005103A2?OpenDocument
As far as the noise that you are seeing, you
may be experiencing a ground loop. This is caused by your instrument being at a
slightly different (perhaps in mV) ground potential. Look at these two articles
to determine if this applies to you and some steps that you can take to resolve
this noise.
Avoiding Ground Loops
http://zone.ni.com/devzone/cda/tut/p/id/3035
Inconsistent Readings Measuring
Ground-Referenced Signal in Referenced Single Ended (RSE) Mode
http://digital.ni.com/public.nsf/allkb/DE02FF8099B8427686256297005224B2?OpenDocument
02-14-2008 04:09 AM
02-14-2008 04:47 PM