KM10 Pressure Transducer Connection via NI9949 RJ50 Terminal Block to NI9237
Good Day,
I am new and forgive me for any misuse of nomenclature. My goal is to measure the pressure within a chamber from a 3 wire KM10 pressure transducer. I am using an RJ50 female screw terminal block (Serial # 196773B-01) is used to connect the 3 wires. The red I have connected to the excitation port, which is the #6 location of the terminal block, the black I have connected to the negative analog input port which is the #3 location of the terminal block and the white I have connected to the Positive analog input signal connection which is the #2 location of the terminal block. There is also a bare unshielded wire that is connected to the #7 position on the terminal block. I don't know where that would be wired specifically. This is all connected to the NI 9237 Full bridge analog input DAQ though a Rj50 cable. When wired as shown in the attachments and scaling the pressure values to 0.5mv/v at 0 psi and 4.5mv/v at 100 psi along with providing an execution voltage of 5v there is a wrong pressure value and the pressure transducer is unresponsive under load. Am I wiring this correctly or will I need to use a quarter bridge setup? Any guidance on how to properly configure a pressure transducer and if the physical setup is correct will be much appreciated. Thank you
(1,087 Views)
I don't think KM10 is a bridge-based pressure transducer. NI-9237 only accepts the input voltage of +/1 25mV/V. With an excitation of 5V, the maximum range is +/-125mV.
KM10 outputs a signal of 0.5 to 6 VDC. It is just a 3-wire voltage pressure transducer. Use a normal voltage input module with an external power supply as the power supply instead.
https://tmesystems.net/
(1,060 Views)
Do you recommend an NI-9201 for this application? As far as connectivity goes would I just directly screw down the white signal wire to one of the screw terminal ports on that device and just setup as usual though the labVIEW DAQ assistant?
Thank you
(1,055 Views)
NI-9201 is a good choice with an adequate input range. Don't forget the external power supply.
(1,038 Views)
Since I was limited on equipment I wasn't able to use the NI-9201 but I was able to use an Arduino and use the analog input features that come along with it. Thank you for the help.
If you are curious to see a wiring schematic or the code setup let me know and I will link it in the thread
(1,007 Views)