Lost precision in a DAQ Vi.
We are measuring a high value resistance using a slow DAQ Vi in LabView 6.1 and it's fine, but when We try to measure the same resistance using a fast DAQ vi (buffered) the value is changed about a 30%. We keep on mind that there is a capacitance value in the lines and we remove that element. We need to know how improve the data adquisition without lost the precision? Thank you.
(2,450 Views)
Does the value increase or decrease by 30%? If the value is increasing then you might be running into charge injection. Charge injection is caused by sampling at high rates with a high-impedance source. The documents linked below discuss this phenomenon and give some solutions. I would look into decreasing the source-impedance by using a unity-gain buffer circuit as explained in the documents below. I hope this helps.
Is Your Data Inaccurate Because of Instrumentation Amplifier Settling Time?
http://zone.ni.com/devzone/devzoneweb.nsf/Opendoc?openagent&F7B07C1AEFAC727B8625686500693572
Using a Unity Gain Buffer (Voltage Follower) with a DAQ Device
http://zone.ni.com/devzone/conceptd.nsf/webmain/CD57A73721E0612586256BAE0055CDD9?opendocument
(2,450 Views)