Maximum current draw
This is the first time I have needed to use analog output. I have 15 pressure sensors that will each draw 4-6 mA at 10V. I want to use the NI 6514 to provide my power to these tranducers which will return a 4 - 20mA signal I will bring into and SCXI 1102 and PXI 6221. Looking at the specification maximum current output is 75mA. Is that per line, per bank or total? Will I need more of these cards or will one card be able to provide enough current?
If the 75mA is per bank I should have no problem, but my thinking is that the loads (sensors) are all in parallel to the common Vs and currents in parallel add. This would mean that if I was using all 32 outputs at +10V, I am looking at less than 2.5mA per channel that could be driven.
Please someone help clarify.
Sincerely,
William.
(3,175 Views)
bumping this message up looking for an answer.
(3,152 Views)
Hello William. Thanks for contacting National Instruments. The maximum current output is 75mA per line. This should meet the specs that you have listed above. Have a great day!
Marni S.
(3,128 Views)