Multifunction DAQ
Measurement Noise NI-9226 vs. NI-9219 for PT1000 sensor
I am looking for a DAQ solution for measuring 2 PT1000 sensors. My measurements will be done in a temperature stable (20 degrees C) environment and will take about 10s each. I would like to be able to measure temperature changes during these 10s in the order of (sub)mK with a sample rate around 2sps (both sensors). I want to interface the DAQ with matlab, so the cDAQ + USB range is where I looked.
I've found two devices that may give me what I need; NI-9226 (states a 1mK rms noise ~ 3mOhm) and the NI-9219 0.4ppm rms of range. Now my question is:
- How do I express the 9219 noise in mK for comparison? What is the 'range' when I use a PT1000. sensor? Is it simply 1000*0.4ppm = 0.4mOhm (therefore a better noise performance than the 9226?) or am I missing something?
- Am I overlooking a device that can be used or is the 9219 my best option for noise performance?
I hope someone can clarify this for me.
Thank you and best regards,
Jasper
(2,822 Views)
Accepted by topic author J.D.Keuning
Hello Jasper,
So far your assumptions are (nearly) correct.
The actual range of the 9219 for PT1000 is 5.05kOhm (see page 8 of the specifications).
So the noise specification is 5.05kOhm*0.4ppm = 2mOhm ~0.7mK
So the 9219 is still the best in terms of noise, but only by a slight margin .
To my knowledge those two are the only devices which can be used.
Andreas
Andreas
CLA, CTA
(2,786 Views)
Hi Andreas,
Thank you for your reply and pointing me to the actual range. I've gone and bought the 9219, so if I missed some other critical information, I'll find out soon!
Jasper
(2,764 Views)