Irene,
The way you are doing acquisition and using the ‘split signals.vi’ is correct. The best way for you to keep track of which channels are being graphed is to rename the channels in the DAQ assistant, these names will then follow on to the graph. I have attached a word document demonstrating this with further explanation.
One note is that the 9205 can only do analog input. ( your document showed analog output as well ). If you are using another module to do output, depending on what your DAQ assistant is set for (single point, continuous, etc), you may need to input arrays to the 'merge signals.vi' and not just single points.
Message Edited by David L. on 04-06-2007 10:47 AM
04-03-2008 09:14 PM
Hi David,
I have a question about 9205 and 9215
I am trying to get reading by that modules from K-type thermocouple and results I am getting look abnormal. there is a jumps in readins +-200 dergees C, if I am trying to measer voltage (6-7V) 3rd digit is not stable (like 6.2 stays same and 3rd digit is changing all the time significly). Could you suggest what's wrong?
best regards,
Anton
04-04-2008
05:15 PM
- last edited on
09-25-2025
08:26 AM
by
Content Cleaner
Hi Anton,
What you described about the error when measuring 6-7V seems to be within specification of both devices. Refer to the manuals for the NI-9215 and NI-9205 for details about the accuracy specifications.
As for the temperature measurements, these particular devices don't have built-in CJC. What value are you using for the CJC when configuring your task? Can you also provide some details about how you are configuring the thermocouple task?
04-05-2008 11:03 PM
Hi, I am trying to get reading by M&A 4.2 by creating task for thermocouple readings. I am using K-type thermocouple so I set that parameter in task. I know that there is no CJC in that module, that's why I put 24C (appr. room tempr.). but I am not expiriencing wrong reading but fluctuations within 300-400 degrees, regular multimiter shows 1030 C from that thermocouple, but my module is jumping within 850-1200 C. We got that module after consultation with NI rep. we dont need 100% reading, +-2-3 degrees acceptable for our application, but here we have something strange.
thank you,
Anton
04-07-2008 06:02 PM
04-18-2008 03:54 PM - edited 04-18-2008 03:55 PM
04-18-2008 03:59 PM - edited 04-18-2008 04:00 PM
Take a look at one more picture attached. As you can see there is steady signal on line 0 (channel 0), but rest 7 have kind of sinus signals. still need help.
thank you,
Anton
04-22-2008
10:51 AM
- last edited on
09-25-2025
08:26 AM
by
Content Cleaner
Hello Anton,
The floating signal behavior you're seeing on non-connected channels is expected behavior. When you don't have anything connected to these channels the voltages are floating. Since the inputs are all multiplexed through one analog to digital converter (ADC) when you connect a 5V signal to one input the other unconnected channels will have a similar reading due to residual charge on the ADC. This charge on the ADC can't be dissipated when there is nothing connected to the line. If you wanted to read 5V on one channel and 0V on the other channel then you'd need to ground all of the other channels.
Here are a couple KnowlegeBase articles that talk about this behavior: Troubleshooting Unexpected Voltages, Floating, or Crosstalk on Analog Input Channels and Incorrect Readings when Scanning Unconnected or Open Channels (Option 3 in the second KB is the one I mentioned above).
I hope this helps, and have a great day!