Hello Massimo,
Please see https://forums.ni.com/t5/Multifunction-DAQ/AI-GND-vs-AI-SENSE-for-NRSE-inputs/m-p/628660#M35888
I agree with Chris_D that the bias resistor connecting AI SENSE to AI GND will not affect the GRS. Another NI document that is worth while reading is the Field Wiring and Noise Considerations for Analog Signals (https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...). It gives a thorough explanation of the different signal sources, and measurement configurations. To reiterate the configuration for your application, you should set all the channel switches set to SE, and the RSE/NRSE set to NRSE.
Since there is no access to the AI GND or AI SENSE on the connector pane of the BNC 2090, connecting the bias resistors is difficult.
Connecting Signals to the BNC 2090 KB (https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000001DdSnSAK&l=en-US) explains how to connect these bias resistors, but without having access to a 68-pin break out box you will need to open the BNC 2090 as indicated in the manual and solder the resistors in.
Have a good evening
Chris_K
12-19-2007 04:29 AM
12-20-2007 09:19 AM - edited 12-20-2007 09:20 AM
Hello Massimo
The information for the type of signal source of the signals that you are using to take measurements can be found in their respective manuals.
The following image details the difference between an ground referenced signal, and a floating signal.
A battery or thermocouple would be a floating signal because it doesn’t explicitly reference building ground. A sensor where the negative terminal is connected to ground would be a ground referenced signal. You could check continuity between the negative terminal of your sensor and ground and if there is continuity then it is a GRS, if not, then it is floating.
Have a good day
Chris_K_