Hi Dear all, The nominal ranges of NI 9205 are ±10V, ±5V, ±1V, ±0.2V. Will I be able to measure something smaller than ±0.2, for example 30mV?

I am using labview.

 

Thanks and regards, Y

0 Kudos
Message 1 of 7
(4,987 Views)

Hey Yunsan,

 

Those programmable ranges are there to provide you with better resolution if you know the maximum/minimum voltage will be inside ±0.2V (for instance).

 

To demonstrate this, the ADC on the 9205 has 16 bits of resolution.  At ±10V, the bin size for the device would be 20V/2^16 = 305 microvolts.  This means that the device would be able to measure changes in increments of 305 microvolts.  Now if the maximum voltage range is set to ±0.2V, the bin size would change to 0.4V/2^16 = 6.1 microvolts, which can be useful as long as you know that your measurement will not go outside of this range.

 

Another thing to consider is the Absolute Accuracy of the device at those programmable levels.  According to the Operating Instructions on page 22, the AI Accuracy of the device changes with this setting as well.  For example, at ±10V the Absolute Accuracy at Full Scale 6.23 mV, which means that while the given digital value will fall within a bin size of 305 microvolts, the actual reading of the device that falls within one of those bins could be up to ±6.23 mV off.  Now if set to ±0.2V, the accuracy would be 174 microvolts.  This means that again, the bin size is 6.1 microvolts but the actual reading could be up to 174 microvolts off.  Information on other things to consider such as noise and sensitivity can be found in the document linked above.

 

So to answer your question, yes you can measure 30mV with either the setting of ±10V or ±0.2V.  The difference in these settings for that example would be (taking just accuracy into consideration) that for ±10V your 30mV input would measure within the range of 30mV ± 6.23mV and for ±0.2V it would measure within the range of 30mV ± 0.174mV. 

 

A suggestion would then be to use the smallest nominal range that you know you can get away with.

Regards,
Jake G.
National Instruments
Applications Engineer
Message 2 of 7
(4,970 Views)

Thank you very much!

0 Kudos
Message 3 of 7
(4,965 Views)

Sorry to hijack the post but can one of you tell me how to programmatically set the range?

0 Kudos
Message 4 of 7
(4,734 Views)

Hi DutchTzu13,

 

If you're using LabVIEW and NI-DAQmx, you can programmatically set the range with a DAQmx Channel property node and the AI.Max and AI.Min properties.

 

Brad

---
Brad Keryan
NI R&D
0 Kudos
Message 5 of 7
(4,725 Views)

Awesome, thanks!

0 Kudos
Message 6 of 7
(4,714 Views)

Between the Absolute Accuracy and the Sensitivity, I'm a little confused.  Say I'm measuring a voltage that is 10uV, and the sensitivity is 4uV at the +/-0.2V range and it has a 174uV Absolute Accuracy (and also 10uV Typical Noise).  What kind of range of measurements, minimum and maximum, am I likely to see taking all this into consideration?