Hello!
I have not been able to find an external calibration procedure for the NI 9223 Analog Imput Module. I have Calibration Executive 3.4.1 and it is not listed, and I have searched the manual lists for a manual procedure and cannot locate one. Does an external calibration procedure exixt for the NI 9223?
Thank you!
11-14-2011
12:53 PM
- last edited on
02-16-2025
05:18 PM
by
Content Cleaner
Hi Phillip, calibration for the NI 9223 is not necessary according to the article "Equipment Required to Calibrate NI Measurement Hardware". If you believe that your hardware needs to be repair then please visit the NI Repair Services page for further assistance. I hope this helps
11-14-2011
01:26 PM
- last edited on
02-16-2025
05:21 PM
by
Content Cleaner
Actually, calibration is recommended/required every 1 year (see the Calibration Interval spec).
NI is still in the process of creating an external calibration procedure for the 9223. I'm not sure of the timeline for the availability of an official document. However, in the meantime NI does offer factory calibration of the 9223.
Do you have calibration requriements beyond what is offered by the factory calibration?
Best Regards,