During my internship I got the order to build a prototype for component tests.
So I need to automatically increase the supply current for the components until a certain level (probably around 5 to 6 mA) is reached. Then I'll get a digital input as feedback from the component which I want to use to stop the increase. The same thing has to be done by decreasing the current to detect not just the level but the hysteresis as well. I need to report both values for each component.
The increase or decrease steps should be around 0,05 to 0,1mA.
I already have the cDAQ 9188, therefore I want to use the NI 9265. But I'm not quite sure if the resolution (16-bit) is suitable for me.
The minimum step which is possible should be
(20*10^-3) / (2^16) = 305,17.... nA.
Am I right?
I tried some examples but everything I found was a 1mA step as the minimum value.
Thanks for help and advice in advance.
Cheers,
Pascal
09-17-2015 10:53 AM
Hey Pascal
according to the data sheet:
NI 9265 Datasheet - 374067a_02.pdf
http://www.ni.com/pdf/manuals/374067a_02.pdf
the step size for the 9265 0-20mA current output module is
20mA/2^16 = 0,00030517mA = 0,30517uA
But you have to take the accuracy (gain and offset) as well as the temparature into account.
Regards, Stephan