Hello everybody,

I would like to know if it is possible to perform NI-9266 calibration with cDAQ-9185, 4-slots Ethernet châssis. I saw that cDAQ-9178, 8-slots USB châssis, is recommended for that.

Thank is advance for your replies,

Best regards,

James

0 Kudos
Message 1 of 3
(1,124 Views)

If you are referring to the NI 9266 Calibration Procedure, I don't see why cDAQ-9185 won't work. You are calibrating the module but not the chassis. Both Ethernet and USB chassis use DAQmx driver.

-------------------------------------------------------
Applications Engineer | TME Systems
https://tmesystems.net/
0 Kudos
Message 2 of 3
(1,078 Views)

Hello all,

Yes Zyong, this is this procedure. It is mentionned on page 4 at paragraph "Initial Setup" at step 3 "Insert the module in slot 8 of the cDAQ-9178 chassis. Leave slots 1 through 7 of the cDAQ-9178 chassis empty.".

But in cDAQ-9185 chassis, there are only 4 slots. There is no slot 8.

Which cDAQ-9185 slot should I consider for calibrating my NI-9266 module?

Thanks in advance for your replies

James