09-04-2012 03:44 AM
Hi
I am using NI digitizer PCI 5105. It was working fine but now its shows an error "Error -239020 occured at Self-test for national instruments" and "Internal hardware error occured in MIG hardware. Please contact national instruments support". What might be the reason? Please help me to solve this problem.
09-05-2012
11:05 AM
- last edited on
06-17-2024
01:59 PM
by
Content Cleaner
Hi Paritosh
Typically this error indicates the device has trouble communicating with the onboard FPGA. Do you know what has changed from the time it was working to the time you started seeing this error? To determine if this is a hardware or a software issue, please try these steps to see if they solve the issue-
If none of the above steps solve the issue, you may need to call in on our support line at 866-ASK-MYNI (866-275-6964) and create an RMA request to return the board for repair.
09-05-2012 08:53 PM
Thank you for your reply. I can not reset the device in MAX because it shows the same error (Error -239020 when I press the reset button. I don't think there's any problem with the driver because I tested the device in 3 PCs and it didn't work on any of them. However, when i connect another PCI 5105 board which I have to all these PCs and it works fine. So I suspect it might be a hardware problem. Can i do anything to make sure of it?
09-06-2012 04:15 PM
Hi Paritosh,
When you connect another PCI 5105 board to the PC, is it in the same slot as the not working 5105? If so, this is a pretty good indication of a hardware issue and you can call in on our support line for an RMA. Thanks and have a nice day.
09-07-2012 06:25 AM
I connected the working PCI 5105 in the same slot. So i think it is a hardware problem. I'll contact the support line you've given me. Thanks for the help.