02-15-2007 06:32 PM
What is the reason for having just noise when testing Analog I/O?
How can I do to discard hardware failures?, I’m suspecting hardware failures particularly because without any external wired up, the AO GND (Pin 54) is at ground level but AO GND (Pin 55) has higher potential.
I’m testing using the Test Panel (Measurement & Automation Explorer), Signal generator and Tektronix oscilloscope.
The results are as follow:
Self test passed alright
Digital I/O are fine.
+5V are fine
ANALOG INPUTS
The signal generator as an input means it’s a Ground-Reference signal source therefore the input configurations tested are Differential and NRSE described as follows:
TEST 1:
Analog Input
Channel: Dev0
Mode: On Demand
Input configuration: NRSE
Pin wired AI0 (Pin 68) AISENSE (Pin 62)
TEST 2:
Analog Input
Channel: Dev0
Mode: On Demand
Input configuration: Differential
Pin wired AI0 (Pin 68) AI8 (Pin 34)
TEST 3:
External measurement was made using a Tektronix oscilloscope and showing as a result just noise although I change the output range and voltage.
ANALOG OUTPUT
AO0/1
Mode: DC Voltage
02-20-2007
04:19 AM
- last edited on
07-22-2025
03:10 PM
by
Content Cleaner
Hi,
Thanks for posting your query on the National Instruments discussion forum. I’m wondering if you are using a shielded cable in your measurement setup. The PCI 6036E Multifunction DAQ card doesn’t have any onboard filtering so adding filters for each channel will help increase the signal to noise ratio. Ideally you would add filters as close to the signal source as possible, as this is where it is most effective. Have you used any external signal conditioning accessory like NI SCXI?
This article is great source of additional information about signal conditioning
I hope this helps!
Kirtesh Mistry
National Instruments UK & Ireland
02-20-2007 06:52 PM
Thank you very much for your answer and link.
I'm testing with twisted pair of 5cm length because I just want to check the DAQ card before going to the full implementation. Also I was following the schemes suggested in the link that you have recommended.
The result is not a “noisy reading” it is just noise that doesn't exhibit any change when I change the signal generator or the range setting in the MAX options.
I have tested also the self diagnostic using internal virtual channels and the result is the same.
Could this behavior be the result of hardware failure?
02-21-2007
03:56 AM
- last edited on
07-22-2025
03:11 PM
by
Content Cleaner
Hi,
Can you verify external electrical readings being supplied or being read from the card, using a multimeter? Have you had the card in working order / condition at any point prior to this issue that you have reported? I'm not sure what driver version you may have installed but just to let you know, you can obtain the latest driver (free download) for the card from www.ni.com/downloads > drivers and updates link.
Productline - multifunction daq
Software - NI DAQmx
Version - 8.3.1
Operating System - ?
Essentially - it is better to exhaust all avenues before concluding that the card is not working because of hardware failure. - save you the hastle of going through RMA procedure / purchase. Also knowing this information would help us with any 'repair' and return requests.
I hope this helps!
Kind Regards,
Kirtesh Mistry
Applications Engineer
National Instruments UK & Ireland
02-22-2007 09:18 AM
Thank you very much for helping me “to exhaust all avenues”, always it’s something else to try.
I have yet installed the last driver version (NI-DAQmx 8.3.1f0)
Summary of results:
Measurements media:
The board was used before by different research groups therefore I have yet consulted to other owners about the functioning of the DAQ card and we all agree that must be some physical failures that require the knowledge of NI engineers.
Therefore I would like to know what the next step is in order to solve the problems.
Thank you very much again for your collaboration
02-26-2007
10:34 AM
- last edited on
07-22-2025
03:11 PM
by
Content Cleaner
02-28-2007 06:26 PM
03-05-2007 03:21 AM