Hello Mozzi,
The fact that when you connect the probe ground to the board, it works, makes it quite obviously a grounding issue. Take a look at the tutorial Field Wiring and Noise Considerations for Analog Signals. Pay particular attention to the Table I, which summarizes this different types of terminal configurations. Your signals may be improperly routed to the DAQ board. You might want to "Ohm out" each of your traces to make sure you have good connections
On the other hand, the measurement results you've described make it sound like a phenomenon known as ghosting. Take a look at the KB How Do I Eliminate Ghosting From My Measurements? and the tutorials linked to from it. This may also provide some insight into the problem. If it is ghosting causing the problem, you should be able to read from each channel individually and get the correct signal. Then if you were to scan through all the channels in succession, you would see the ghosting return.
I would first consider the possible ground issues that might be involved. I don't think it has anything to do with ghosting, but I just figured I would mention it. Unfortunately, we will not be able to directly support your custom interface board, so other than the suggestions I've made, there's not much else that I can do to help. Hopefully the information I've provided will be helpful.
Best regards,