Noisy Signal from USB 6211
Hello All,
I'm very new here and need help.
I just got new USB 6211 and used to get signal from lock-in amplifier.
The signal seems to be very strange. To check it, I applied only 5v from the lock-in and the signal is so weird. See the attached file.
it suppose to be 5v output, but as you can see it is noisy at some interval, goes to zero at other intervals, and solid 5v sometimes in a repeatable systematic way!
Can you please guide me through this.
thanks
Message 1 of 11
(5,149 Views)
Does the lock-in amplifier take some time to 'lock in', and then after a period of stability, reset so that it can get a new measurement? It's my guess that's what's happening for you.
If you provide some information about the amplifier (model number, or preferably a link to/upload of the manual) then we could look closer.
GCentral
Message 2 of 11
(5,128 Views)
the lock-in is the standard one from Stanford research ( SR830 )
however, I believe it is not lock-in issue. Cus i got similar response with 2v battery
Message 3 of 11
(5,125 Views)
Sorry - to confirm: you connected a 2V battery to the 6211 (presumably as an analog input) and the behavior was similar? At what sampling rate - I note the graph you posted a picture of shows THz sampling and the 6211 only supports up to 125kHz. The SR830 only claims to be able to reach around 100kHz and has outputs to Serial and GPIB.
What are you connecting to what, and where did the THz graph come from (or is the scale just completely wrong?) I'm pretty confused...
Edit: I see this in the manual for a possible analog input to 6211:
AUX OUT 1-4 (D/A Outputs) These are auxiliary analog outputs. The range is -10.5V to +10.5V and the resolution is 1 mV. The output impedance is <1Ω and the output cur-rent is limited to 10 mA. These outputs may be programmed from the front panel ([Aux Out])or via the computer interfaces.
X and Y The X and Y lock-in outputs are always available at these connectors. The bandwidth of these outputs is 100 kHz. A full scale input signal will generate ±10V at these outputs. The output impedance is <1Ω and the output current is limited to 10 mA.
GCentral
Message 4 of 11
(5,120 Views)
Thanks for the reply.
please ignore the THz part. Im just using this labview software to view the analog data coming from the lock-in
Message 5 of 11
(5,116 Views)
Can you show a screenshot of the 2V battery output? I'm surprised that it's the same, but if so, it makes the amplifier a less likely suspect (as you described).
GCentral
Message 6 of 11
(5,111 Views)
Thanks.
i dont have the 2v picture now, i may update it later.
i was reading other posts and i suspect there might be issue in the connection.
i made the connection between Al0 and AI8. Some seems to suggest that there might be grounding issue. Do i have to connect them with the AI GND?
Message 7 of 11
(5,109 Views)
If you're using the Aux Output, Channel 1 or Channel 2 outputs, they should be BNC connectors. If you're splitting the cable to two outputs, then wiring both to the 6211 in a differential configuration, I'd have expected you to be OK...
If you're splitting it but only wiring the core, positive output, then yes - you might have a problem with your ground. If you connect all of the grounds together and then connect that to AI GND, then you can make single-ended measurements of the positive parts of the BNC output, but I don't know for sure that the 830 has shared negative/ground outputs on each terminal, so take care if it's more than one output from amplifier/input to 6211.
GCentral
Message 8 of 11
(5,106 Views)
Yes, i have splitted  the BNC into two cables. One to AI0and the other to AI8.
i think the lock in amlifier is grounded
everything seems correct to me but dont know where is this issue coming from!!
any advise on what else i can do to test the device?
Message 9 of 11
(5,101 Views)
I think to get further we need to look at the code you're using to acquire the data in the graph.
If you're sure it isn't a wiring problem, and you're reading a 5V constant output from the BNC Aux Out (1-4) channel on the amplifier (you could check this with a separate standalone DMM perhaps), then either you have some peculiar repeating noise (I note the drops are at intervals of around 60 'ps', but I don't know what the real scale is, so it's hard to draw anything from that) or the code has some strange problem.
Or perhaps the 6211 is broken, but I'd like to continue to doubt that for at least a little while longer.
GCentral
Message 10 of 11
(5,096 Views)
