arj- I think you slipped in the units
0.2/65536 ~ 0.000003 = 3uV (microvolts).
You may be picking up external noise if your signal is still drifting or not what you expect it to be.
01-04-2008 12:14 PM
01-04-2008
01:31 PM
- last edited on
11-10-2025
11:52 AM
by
Content Cleaner
Hi arj,
There’s a difference between accuracy and resolution. Resolution is the voltage range divided by 2 to the bits of resolution, whereas accuracy is the specification you use to determine the overall uncertainty of your measurement. Absolute accuracy specifications apply only to a successfully calibrated device and within a specific temperature range. The considerations for accuracy are better described by this Developer Zone document: Accuracy.
So, some questions that may help guide you to a successful reading include:
1. 1. When did you last calibrate (internally or externally) your card?
2. 2. What is the environment like where you have your measurement system?
3. 3. What is your input signal? There may be other noise considerations to consider depending on what it’s like. (See Accuracy.)
4. 4. Do you get a stable measurement when measuring something very stable, such as a battery?
The Differences Between Accuracy and Bit Resolution
The links I’ve included in the body of the message are very helpful resources, but feel free to repost if you have further questions.
Message Edited by Mark E on 01-04-2008 01:33 PM