Here are the results from the test:
7/2/2008 12:55:07 PM
Results saved to: F:\Diagnostic Results2.txt
Selected Device: Dev2
Device Type: PCI-6071E
Serial Number: 1237207
Device Support: (PASS)
NI-DAQmx Version: 8.0 (PASS)
Device Reset: (PASS)
The device was last self-calibrated on: 10/17/2006 3:36:39 PM
Running Calibration ... FAILED for the following reason:
Error -200545 occurred at DAQmx Self Calibrate.vi
Possible reason(s):
Self-calibration failed. The self-calibration date has not changed.
Disconnect the device from external signals, as they might introduce noise. Externally calibrate the device to recalibrate the onboard voltage reference that is used for self-calibration.
-------------------- Testing Counters... --------------------
Testing CTR Buffered Pulse Generation... (PASS)
Testing CTR Single Pulse Generation... (PASS)
-------------------- Testing Analog Input... --------------------
Testing AI Ground Reference... (FAIL)
************ Test FAILED for the following reason ************
Analog Input was not within tolerance for ground reference. Expected to read: 0.000000 Actually read: -10.000000
*******************************************************
Testing AI Voltage Reference... (FAIL)
************ Test FAILED for the following reason ************
Analog Input was not within tolerance for reference voltage. Expected to read: 5.000000 Actually read: -10.000000
*******************************************************
Testing AI Finite Sample Clock... (FAIL)
************ Test FAILED for the following reason ************
Count of sample and convert clock pulses does not match expected number. Frequency: 20000.00 Samples: 10.00 Channel List: ai0:4
*******************************************************
Testing AI Finite Sample Clock Frequency... (PASS)
-------------------- Testing Analog Output... --------------------
Testing AO Finite Sample Clock... (PASS)
Testing AO Finite Sample Clock Frequency... (PASS)
Testing AO Single Point... (FAIL)
************ Test FAILED for the following reason ************
This test failed to output 9.000000 Volts on Dev2/ao0. The actual value written was -10.000000.
*******************************************************
Testing AO Accuracy... (FAIL)
************ Test FAILED for the following reason ************
This test failed to generate a voltage waveform with an error of less than 0.385802 Volts on Dev2/ao0. The actual average error was 10.000000 Volts.
*******************************************************
-------------------- Testing Digital IO... --------------------
Testing DIO Lines... (PASS)
Testing DIO Ports... (PASS)
----- DIAGNOSTIC TEST COMPLETE -----
Thanks,
Matt