Hi Stoico,
THe M Series cards ( PCI 6259/6254 belong to M series) offer only Bipolar Mode.
There were some E series cards ( For Ex: PCI 6070E /6071E) in which you could choose between Unipolar and Biplolar modes for AI channels
But the M Series Cards have better Absolute accuracy in comparison with E series cards
Here is the link to the specifications of M series PCI 625x Family: https://www.ni.com/docs/en-US/bundle/371291g/resource/371291g.pdf
And here is the link to the E series Card's specifications: https://www.ni.com/docs/en-US/bundle/370725c/resource/370725c.pdf
This should help you compare between both and choose better
Hope this helps
Regards,
Dev
			
    
	
		
		
		12-18-2007
	
		
		05:45 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 - last edited on 
    
	
		
		
		08-26-2025
	
		
		11:39 AM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 by 
				
		
		
			Content Cleaner
		
		
		
		
		
		
		
		
	
			
		
Hi Stoico,
I just wanted to add one more piece of information for you to consider. Dev was right when he said that there were E series cards that supported unipolar ranges and that M series are more accurate (along with other benefits) than E series. The good news is that you can still have the accuracy of the M series and also get unipolar measurements with them.
To measure unipolar signals with M series cards, you choose the smaller range you want and provide an offset. This is done with the non-referenced single-ended (NRSE) mode. The M Series device measures the voltage of an AI signal relative to one of the AI SENSE or AI SENSE 2 inputs. This would be your offset voltage. See the M Series User Manual (chapter 3 for signal descriptions and chapter 4 for ground-reference settings).
I hope this helps you make the right selection.
06-26-2012 12:01 PM
Am I correct in assuming that the M series cannot be put in a unipolar mode from 0 to 20V, as this would exceed the 11V maximum difference between AI+ and AI GND?
I would like to use the ADC capabilities for a signal that varies from 0-18V. I'm wondering if there is a way to configure the input references so that I won't have to do any biasing on the signal. Any suggestions or workarounds would be appreciated. Thanks,
Joel
06-27-2012 04:16 PM
Hi Joel,
Yes, you are correct. You can do a unipolar measurement of 0 to 10V but not 0 to 20V. If you want 20V range you might want to do differential measurement of -10V...+10V.