PCI-6602 edge counting with pause trigger has resource conflict
I have a customer with a PCI-6602 counter device. He's trying to use our software to set up several counters for unbuffered edge counting with a pause trigger (that's NI-speak for "gate") so that he can control when the counters count. He wants to assign arbitrary PFI's to the counter sources and gates.
My code (abbreviated- rather complicated code):
DAQmxCreateCICountEdgesChan(GetTaskHandle(), getChannelName().c_str(), NULL, edge, initialCount, dir)
DAQmxSetCICountEdgesTerm(GetTaskHandle(), getChannelName().c_str(), SourcePin.c_str());
DAQmxSetPauseTrigType(h, DAQmx_Val_DigLvl)
DAQmxSetDigLvlPauseTrigSrc(h, triggerSrc.c_str())
int32 trigger_level = edgeslopewhen ? DAQmx_Val_High : DAQmx_Val_Low;
DAQmxSetDigLvlPauseTrigWhen(h, trigger_level))
In his set-up this works fine for four counters, but the fifth gives an error -89137 when the task is started (or committed). Here is some debugging output that shows the connections he wants to make, and the resulting error information:
  dev: Dev4, srcTerm: /Dev4/PFI12, gat: /Dev4/PFI20, ctrNum: 0
  dev: Dev4, srcTerm: /Dev4/PFI13, gat: /Dev4/PFI21, ctrNum: 1
  dev: Dev4, srcTerm: /Dev4/PFI14, gat: /Dev4/PFI22, ctrNum: 2
  dev: Dev4, srcTerm: /Dev4/PFI15, gat: /Dev4/PFI23, ctrNum: 3
  dev: Dev4, srcTerm: /Dev4/PFI8, gat: /Dev4/PFI16, ctrNum: 4
  NI-DAQmx driver error while calling function DAQmxStartTask; device Dev4:-89137: Specified route cannot be satisfied, because it requires resources that are currently in use by another route.
Extended info:Specified route cannot be satisfied, because it requires resources that are currently in use by another route.
Property: DAQmx_CI_CountEdges_Term
Property: DAQmx_CI_CountEdges_ActiveEdge
Source Device: Dev4
Source Terminal: PFI8
Required Resources in Use by
Task Name: Dev4_CTR_4
Source Device: Dev4
Source Terminal: PFI16
Destination Device: Dev4
Destination Terminal: Ctr4Gate
Status Code: -89137
Seems like the PFI's were never assigned to anything, so they should be available. Is there a default assignment? Do I need to disconnect something in order to connect it to the gate/source?
My apologies for the long and complicated posting; I couldn't figure out a simpler way to post it while including all the necessary info. Any help very much appreciated.
Windows Vista
NI-DAQmx 9.3.0f2
PCI-6602
John Weeks
WaveMetrics, Inc.
Phone (503) 620-3001
Fax (503) 620-6754
www.wavemetrics.com
(3,720 Views)
Hi John,
It appears the issue is that the route of PFI8 to Ctr4 Source uses TIO/Internalline8 to route the signal which PFI16 to Ctr4 Gate uses as well.  You can change the terminal from PFI8 to "Ctr7InternalOutput" and the error should go away as this terminal does not utilize the same line.  Let me know if that works for you, and/or if you have any further questions.
Ben N.
Applications Engineering
ni.com/support
(3,701 Views)
Thanks, Ben. That explains the error. Unfortunately, the customer simply wants a bunch of independent counters, each with a gate. So he needs to come up with compatible pairs of PFI's to use for source/gate pairs.
How would I find out about the conflict in the internal route, except by trial-and-error? The MAX Device Routes table for the 6602 device show all PFI-to-CtrNSource and PFI-to-CtrNGate routes as being direct routes.
John Weeks
WaveMetrics, Inc.
Fax (503) 620-6754
www.wavemetrics.com
(3,696 Views)
Accepted by topic author WM_John_Weeks
John -
I assume you are familiar with Device Routes which can be tab-selected in MAX for a particular device.  In that table of routes you will see white (blank), yellow and green boxes tiled throughout the matrix of the device's terminals.  The blank is a no-route, yellow is indirect and green is direct.  It's running into the indirect routes that is going to cause a potential resource use issue.  If you hover your mouse cursor over the Yellow or Green box on the Device Route Chart, you can Shift+Click and it will popup the internal device route.
In your customer's instance, PFI 8 and PFI 16 were both using Internal line 8 for the indirect route to both the Ctr4 Source and Gate respectively.
Ben N.
ni.com/support
(3,693 Views)
Ah, thanks Ben. I was missing the shift-click part.
What I see in MAX is that the blocks connecting PFI's to counter source/gate are all green, but the shift-click information shows that the internal routes use something called TIO0/InternalLineN. Indeed the routes for PFI8 and 16 both use InternalLine0.
John Weeks
WaveMetrics, Inc.
Fax (503) 620-6754
www.wavemetrics.com
(3,691 Views)