PXI-5102 Calibration

The 5102 user's manual states that calibration for the board must be done at
the factory. Is this still the case? Have any customer-level calibration
procedures for the unit been released?

Thanks,
Chris Lincoln
ULTRAX, Inc.
0 Kudos
Message 1 of 3
(3,193 Views)
NI-Scope 1.6 was recently released that includes complete calibration
procedures for all NI scope products (5102, 5911, 5112).

The 5102 supports a self calibration procedure with NI-Scope 1.6 that
calibrates the timing of RIS acquisitions. The vertical offset and gain may
be calibrated with an external calibration procedure on this board.

NI-Scope 1.6 may be downloaded from http://ni.com/idnet

Michael Enz

Chris Lincoln wrote in message
news:398f0913@newsgroups.ni.com...
> The 5102 user's manual states that calibration for the board must be done
at
> the factory. Is this still the case? Have any customer-level calibration
> procedures for the unit been released?
>
> Thanks,
> Chris Lincoln
> ULTRAX, Inc.
>
>
0 Kudos
Message 2 of 3
(3,193 Views)
Thanks much! ( and just in time! )

"Michael Enz" wrote in message
news:39901a87@newsgroups.ni.com...
> NI-Scope 1.6 was recently released that includes complete calibration
> procedures for all NI scope products (5102, 5911, 5112).
>
> The 5102 supports a self calibration procedure with NI-Scope 1.6 that
> calibrates the timing of RIS acquisitions. The vertical offset and gain
may
> be calibrated with an external calibration procedure on this board.
>
> NI-Scope 1.6 may be downloaded from http://ni.com/idnet
>
> Michael Enz
>
> Chris Lincoln wrote in message
> news:398f0913@newsgroups.ni.com...
> > The 5102 user's manual states that calibration for the board must be
done
> at
> > the factory.
Is this still the case? Have any customer-level
calibration
> > procedures for the unit been released?
> >
> > Thanks,
> > Chris Lincoln
> > ULTRAX, Inc.
> >
> >
>
>