Hey Olli,
Pseduodifferential is preferable to a single-ended measurement scheme; here's some information from our website about it:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019YuUSAU&l=en-US
Of course, it is not as desireable as a differential connection, but for your application it looks like it might be the best approach. Note that for pseudodifferential you want to have a small resistance between the negative input and ground; you can see more about that in the link above.
10-18-2012 09:27 AM
Hello Daniel,
thanks for the reply! When I take a look into the manual for the PXI-6115/6120, the picture 4-3 shows a common mode voltage between AI- and AIGND. Therefore I thought to connect them together without any higher resistance than the cable in order to have a proper reference to GND. In fact the manual says not, how to connect the Low signal point with the AIGND...or maybe I did not see this correctly? Or do I have to interpretate the sign for the common voltage (Vcm) as a small resistor? When I take a look at the non-referenced connection in picture 4-5 instead it looks quite clear for me. This was my original connection with the floating transducer signals
In what range should the ressisor value be in my AI0 and AI1 connections? Are there some rules of thumb or should I do an approximation from higher to lower resistance values with some static measurements?
Greetings!
Olli
10-18-2012 09:31 AM
ah yes there is a value given in the link, that you sent... 50 ohms.
I'll try this (after reading it more carefuly ;-))
thanks again!