PXI 6221 Maximum Digital Input Sample Frequency
I need some help with the specs for high speed digital input operation of a PXI-6221 in a PXI chassis running from a PCI-Express MXI controller.
I am looking at the 622X Specifications Sheet trying to determine the maximum digital input scan rate that can be achieved.  In the section on Digital I/O/PFI the DI Sample Clock is listed as 0-1MHz.  Is that the maximum for an externally applied signal or can I run at a faster rate if using clock generated internally by a counter?
Why are Port 0 P0.<0,1> specified differently under waveform characteristics?  Do these two pins work differently than the rest of Port 0 on the 6221 for digital input or is that listing only for high speed digital output operation?  The port sample size is listed as up to 8 bits for a 6221, this implies that all 8 bits of Port 0 can be used as high speed digital inputs.
(3,924 Views)
Hello mvr,
Let me try to answer your questions in order:
"In the section on Digital I/O/PFI the DI Sample Clock is listed as 0-1MHz.  Is that the maximum for an externally applied signal or can I run at a faster rate if using clock generated internally by a counter?"
Yes, 1MHz is the maximum sample rate for Digital Acquisition on this device regardless of whether the clock source is internal or external. You cannot attain a sample rate faster than 1MHz by using a counter.
Why are Port 0 P0.<0,1> specified differently under waveform characteristics?"
Port 0 P0.<0,1> are specified differently for the 37-pin NI PCI-6221. This device
(which is only available for the PCI bus) only has 2 clocked digital lines, as opposed to the 68-pin NI 6221 which has 8 clocked digital lines and is available for both the PXI and PCI interfaces. I hope this information is helpful. Let me know if anything is unclear.
Matt Anderson
Hardware Services Marketing Manager
(3,905 Views)
Thanks for the reply Matt.
The NI 622X specification document on page 6 groups the NI-6221(68 pin) with the PCI-6221 (37 pin) in the waveform characteristics section.  Kind of confusing as it looks like it should be grouped with the NI6225.  Thanks for clearing that up I really appreciate it.
Message Edited by mvr on 04-13-2007 08:22 AM
(3,888 Views)
Hello mvr,
You are absolutely correct, that document does appear to be confusing.
This was reported to R&D (# 48F95IAH) for further investigation.
Let me know if you have any other questions.
Matt Anderson
(3,858 Views)