PXI-6229 Random Voltage on Analog Inputs
Using PXI-6229 and the Sample .NET 2.0 AcqOneVoltageSample.2005 program and I get random values returned even if I short the AI pin to AI GND?
(3,626 Views)
Sounds weird. You can use the DAQ Diagnostics Utility (http://joule.ni.com/nidu/cds/view/p/id/566/lang/de) to check your board.
Christian
(3,616 Views)
I ran diagnostics and everything passes. The Test Panel seems to work fine. The example program counts up when grounded.
I think I figured it out. The default is differential mode, setting RSE mode seems to solve the problem.
(3,609 Views)
Hey,
Sorry for my late response, I had vacation.
You should have a look at the DAQmx help where the different terminal configuration modes are described.
There you also find which mode to use for what signal source (groundend, floating).
Christian
(3,576 Views)