PXI-6289 self calibration failed Error -200545
my PXI-6289 self calibration fails. I checked it  with DAQ Diagnostic Utility 2.0:
-------------------- Testing Analog Input... --------------------
Testing AI Ground Reference...   (PASS)
Testing AI Voltage Reference...   (FAIL)
************ Test FAILED for the following reason ************
Analog Input was not within tolerance for reference voltage.    Expected to read: 6,950000  Actually read: 0,187908
-------------------- Testing Analog Output... --------------------
Testing AO Finite Sample Clock...   (PASS)
Testing AO Finite Sample Clock Frequency...   (PASS)
Testing AO Single Point...   (FAIL)
This test failed to output 1,000000 Volts on PXI1Slot5/ao0.  The actual value written was 0,034417.
I had an issue with 12V supply rail - fuse F5 blew. I changed it, but the same problem still exists. I measured voltages on L1 through L8, are they OK?
L1: 0.7 V
L2: 20 V
L3: -20 V
L4: 38 V
L5: 12 V
L6: -12 V
L7: 20 V
L8: -20 V
What else should I check?
Thank you
Martin
(3,734 Views)
Hello Martin,
Just to be sure - have you disconnected all wires from the card when performing self-calbiration and running the Diagnostic utility?
What are those voltages you've measured?
Jakub Prokeš
NIEE Aplications Engineer
(3,686 Views)
Hi Jakub,
yes, I disconnected all wires. The card is unfortunately malfunctional...
The voltages are measured on small inductors marked L1 to L8. Are the voltages correct? I would like to know if the problem comes from power supply of the board or whether it is bad analog part.
Martin
(3,681 Views)
Hi Martin,
Unfortunately I cannot check these voltages and tell you where the problem is. I suggest you to write an e-mail to techsupport@ni.com with link to this thread and a serial number of the card to start the repair process. Your e-mail will be redirected to proper local support, so you can use your local language.
Jakub Prokeš
(3,668 Views)