Problem Programming 6031E in Labview

I have a 6031E DAQ board installed my system and I am using LabView 5.1.
When I have tried to address a channel over 8 on the board, the VIs give
me an error. It has up to 64 channels of Analog inputs. Is there a special
VI that someone has used to address over 8 channels?

Thank You,
Kurt Zdanio
Ovonyx, Inc.
0 Kudos
Message 1 of 2
(2,874 Views)
Kurt Zdanio wrote in message news:39241f30@newsgroups.ni.com...
>
> I have a 6031E DAQ board installed my system and I am using LabView 5.1.
> When I have tried to address a channel over 8 on the board, the VIs give
> me an error. It has up to 64 channels of Analog inputs. Is there a special
> VI that someone has used to address over 8 channels?
>
> Thank You,
> Kurt Zdanio
> Ovonyx, Inc.

Are any of your lower channels ( ACH 0-7 ) defined as balanced inputs? Most of the E series daq cards have the inputs grouped in
pairs of 8 for balanced inputs. e.g. If you configure ACH 0 for balanced inputs, you use ACH0 for (+) and ACH8 (-). Then, if you
try to configure ACH8 for another input, it'll give you an error ( as the channel is already being
used ).

The next "group" of balanced inputs should start at ACH16 (+) and ACH24 (-).

Or, it is possible that the "accessory" selected for the DAQ card only has ACH0 thru ACH15 defined, in which case you may also get
an error.