Question: Diagnostic Error with 6070E, can it be recalibrated by NI or it must be sent to repair?
I have an PCI-MIO-16E-1 (6070E) card which is now reported unreasonable result with my Summitek Instrument.
runing the self test in MAX reports OK.
but runing self-calibrate function in MAX reports error.
I have use the  DAQ Diagnostic Utility v2.0 to test the device, the result is attached below.
my question is that:
1. what might be the problem with this card, is this card still usable?
2. can it be recalibrated by NI or it must be sent to repair?
Thank you for your help.
PS.  I have an other card with the same model, but the test result is the same. Is it the calibrate methord problem, or they are both damaged.
Diagnostic Test results:
2008-9-19 11:36:39
Results saved to:  E:\Diagnostic Results.txt
Selected Device: Dev1
Device Type: PCI-MIO-16E-1
Serial Number: B929CF
Device Support: (PASS)
NI-DAQmx Version: 8.5 (PASS)
Device Reset: (PASS)
WARNING:  Your device has not been self-calibrated.  For best resuts, it is recommended that you self-calibrate the device.
Running Calibration ...   FAILED for the following reason:
Error -200545 occurred at DAQmx Self Calibrate.vi:1
Possible reason(s):
Measurements: Self-calibration failed.  The self-calibration date has not changed.
Disconnect the device from external signals, as they might introduce noise.  Externally calibrate the device to recalibrate the onboard voltage reference that is used for self-calibration.
-------------------- Testing Counters... --------------------
Testing CTR Buffered Pulse Generation...   (PASS)
Testing CTR Single Pulse Generation...   (PASS)
-------------------- Testing Analog Input... --------------------
Testing AI Ground Reference...   (FAIL)
************ Test FAILED for the following reason ************
Analog Input was not within tolerance for ground reference.  Expected to read: 0.000000  Actually read: 0.073842
Testing AI Voltage Reference...   (FAIL)
Analog Input was not within tolerance for reference voltage.    Expected to read: 5.000000  Actually read: 1.595776
Testing AI Finite Sample Clock...   (PASS)
Testing AI Finite Sample Clock Frequency...   (PASS)
-------------------- Testing Analog Output... --------------------
Testing AO Finite Sample Clock...   (PASS)
Testing AO Finite Sample Clock Frequency...   (PASS)
Testing AO Single Point...   (FAIL)
This test failed to output 9.000000 Volts on Dev1/ao0.  The actual value written was 9.997559.
Testing AO Accuracy...   (FAIL)
This test failed to generate a voltage waveform with an error of less than 0.192901 Volts on Dev1/ao0.  The actual average error was 1.577645 Volts.
-------------------- Testing Digital IO... --------------------
Testing DIO Lines...   (PASS)
Testing DIO Ports...   (PASS)
 ----- DIAGNOSTIC TEST COMPLETE -----
(3,476 Views)
Hey,
Can you first please follow the hints given in this thread: http://forums.ni.com/ni/board/message?board.id=250&message.id=30487&requireLogin=False
Christian
(3,470 Views)
Thank you for your reply.
but I have already install last NI-DAQmx driver and reboot the device, and the problem is the same.
Any other Idea?
(3,461 Views)
Since you have updated the latest DAQmx driver and re-tested this card making use of DAQ Diagnostic Utility v2.0 but with the same result, I suggest you contact with NI support ASAP. It maybe broken.
(3,447 Views)