Range Accuracy and Range Sensitivity
How does one calculate Range Accuracy and Range Sensitivity?
E.g., if full scale voltage is -10V_+10V and it is a 16 bit DAQ card, resolution would be 20/(2^16)=0.305mV
the range accuracy is listed as 4660 microV. How is this calculated?
And what is Range Sensitivity?
Thank you for your help.
Philip
(4,028 Views)
Hello Philip,
I would suggest you take a look at Understanding Instrument Specifications -- How to Make Sense Out of the Jargon.
Accuracy is a measure of the capability of an instrument to faithfully indicate the value of the measured signal.  Sensitivity is a measure of the smallest value of an input signal that an instrument can detect.
The spec you saw for range accuracy with the PCI-6133 is calculated using a formula which takes into account gain error, offset error, and noise uncertainty.  This calculation is shown on page 3 of the NI 6132/6133 Specifications.  Take a look and let me know if you have any questions about the calculation.
I hope this helps!  Please let me know if you have any questions about this. Smiley Happy
(4,013 Views)
Thank you very much. They are excellent references. They answered my questions.
Philip
(3,993 Views)