Hi Brett,
I want to make sure this isn’t a
problem in software first before we deem that this hardware fault. You are
right, the chassis does not really have anything to do with CJC so I tend to
think there is a configuration or driver miscommunication. First off, check to
make sure that in MAX, your chassis that is not working properly is set in the
correct daisy chain index. If you have no other chassis connected to it, make
sure that it is set to daisy chain index 0. This can change if you have set up
another chassis in MAX but it is not running. Also, double check that the
address is set properly.
If this does not solve the issue, delete all chassis under devices and interfaces. Then change the address on your chassis’ front and reconnect it to your DAQ board. Set up your modules as always and make sure the address and daisy chain index are correct.
Finally, if none of the above work, try using one 1102/1303 in different slots other than one and see if you are able to get good readings with CJC enabled. If you cannot and you have the correct daisy chain and address set up, then I’m inclined to think this is a strange chassis defect and would call National Instruments.
02-21-2008 04:31 PM
Patrick -
Thanks for the reply. It is a single chassis configuration, with the the index set to 0 (both via the dip switches and the software setting). Unfortunately I cannot change the address on the chassis right now and reconnect, as we are in the middle of a several-day test (using a spare). Once this is done, I will try forcing it to another address and reconfiguring. In my troubleshooting the other day, I did delete it in MAX and add it in again (and all modules were successfully detected), but this was all using the same address again (0).
We also did try different 1102/1303 combinations in different slots of the same chassis. We hauled over a SCXI-1000 (six slot) as well and directly moved one 1102/1303 combo over, only to see it work there. To double check that we didn't inadvertantly change something, we immediately moved the pair back over to the SCXI-1001 and it still was in this failure mode. It was at this point that we dug out a SCXI-1001 spare and began our test.
Also, we checked the three fuses of the chassis and all were operational. In doing so, we did actually see some poor soldering work, but our techs here indicate this is par for the course with NI chassis they've seen in the past. I don't mean to knock you guys (you have lots of great hardware and a great software suite), but when we're tracing strange hardware behaviors and discover shoddy (albeit functioning) soldering it doesn't exactly make us feel like we're barking up the wrong tree.
Anyways, when our test finishes I'll play around with it a bit more and try different addresses for the chassis.
Thanks,
Brett
02-22-2008 11:31 AM
03-07-2008 01:14 PM
Patrick -
Sorry for the slow response, and thanks for your suggestions. I have been away from this project for a bit but took a look at it today. We are not seeing any funny resistances on the backplane fuses, nor is the daisy chain index set incorrectly. I think we are looking at some weird issue with the chassis.
Where to from here? We fortunately had a spare that works like a champ ... it is also a newer revision of the SCXI-1001. I understand our defective chassis is probably not under warranty, is there any way we can get this thing repaired, or is a new chassis the most economical way of dealing with this?
Thanks,
Brett