TS-T1000,
Could you tell us a little more about your setup?
1. What exactly do you mean when you say “affecting the thermocouple readings”?
2. What NI hardware are you using?
3. Do you have multiple thermocouples attached to your device?
02-25-2008 06:58 PM - edited 02-25-2008 07:01 PM
02-26-2008
12:21 PM
- last edited on
09-28-2025
04:47 PM
by
Content Cleaner
Thanks for the info, however can you be more specific with your answers.
1. What kind of incorrect readings are you observing? Are you seeing a lot of noise on the incoming signal? Or are you just reading incorrect values? If so what are those values?
2. Is the surface to which you re connecting your Thermocouples conductive?
3. Have you been able to measure the current from the surface of your DUT?
4. What exact cDAQ module are you using to measure the temperature of your DUT?
5. What kind of tape did you use to attach your thermocouple to your DUT?
6. Do you get proper readings if you just acquire using one thermocouple?
7. Is there a particular reason why the thermocouple has to be attached to the DUT?
If you are seeing a lot of noise coming in from your Thermocouple it could be a result of common mode voltage. This can be due to the electro-magnetic fields that are created when current flows through your DUT circuitry.
02-26-2008 09:15 PM
02-28-2008
09:09 AM
- last edited on
09-28-2025
04:48 PM
by
Content Cleaner
Hey TS-T1000,
Try using some electrical tape to isolate the thermocouple from the DUT. Since thermocouples require a differential measurement, each lead of the thermocouple produces a specific voltage value based on the temperature being measured. The current coming from your DUT could be affecting these voltage values thus giving you incorrect readings. After you isolate your thermocouple you can then use the built in calibration wizard to account for added insulation.
Can you verify that both your devices are working correctly by reading in a separate voltage source on each device?