06-11-2009 01:53 PM
Hey guys,
I'm using the USB-6251 to acquire data from an output that the device is generating. The output is a chirp signal with starting and ending frequencies that the user can input and it has a fixed amplitude. I am sampling this signal with 4 different input channels, all at the same rates. All this is programmed in LabView. Channels 1,3, and 4 all display the same signal that the output is sending out (which is what I want). However, channel 2's signal becomes smaller as the frequency gets larger.
For instance, if I had the starting and ending frequencies of my chirp signal as 2kHz and 80kHz respectively, the wave sampled in at channel 2 would have the expected amplitude at first, but as the frequncies get larger, the amplitude gets smaller. The opposite happens if I start at 80kHz and end at 2kHz (starts small, ends with the correct amplitude).
So somehow, higher frequencies are being filtered out, but only on channel 2. I made sure all my wires are ok as well. Is there anyway to fix this or is my board just faulty at channel 2?
Thanks,
Jeremy
06-12-2009 10:06 AM
Hi Jeremy,
I would try moving your channel 2 signal to channel 5, so you will be acquiring on channel 1, 3, 4, and 5. If all your channels read correctly in this situation, I would suspect that your channel 2 is faulty. If you believe that the board is broken, you can create a service request at ni.com/ask and be sure to set the Question Type to Returning hardware for repair/replacement/calibration. You can then call 1-866-275-6964 with your service request number that you have created.