Uncalibrated value PCI 6110
I am using a PCI 6110 device, the problem is when I run the test sample the input in DevAI0 is a flat line at 5 V, doesn´t change at all. In MAX, if I select the option to calibrate, before doing the calibration, the uncalibrated value shown is 4.996 approximately. Is there a way to fix that? I don´t have a high precision voltage source mentioned in the calibration procedure.
Message 1 of 8
(4,590 Views)
When you say "flat line" does that mean you are reading the output graphically? Are the 5V readings recorded when you are measuring a 5V precision source? Or is this any time you make a reading?
Also, keep in mind that the calibration you are performing is meant to offset cabling and sensor error. It is not a full calibration and will not be stored in the on board EEPROM as explained here:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000YIMNCA4&l=en-US
Furthermore, it probably doesn't make sense to worry about a difference of 4mV because this is already just about within the accuracy specification of the 6110 at a measuring range of +/-5V:
https://www.ni.com/docs/en-US/bundle/pci-6110-6111-specs/resource/370980a.pdf
The accuracy will likely be worse if the device has not been through a full calibration procedure in the past year. You can perform a full calibration by sending your device to NI or following the instructions here:
https://www.ni.com/docs/en-US/bundle/daq-mio-smio-devices-cal-pro/resource/370937w.pdf
Dale S.
RF Systems Engineer - NI
Message 2 of 8
(4,558 Views)
The flat line at 5V appears without measuring anything else. It also appears in Test Panels in MAX.
Thanks.
Message 3 of 8
(4,552 Views)
Try shorting AI0 to GND and measuring the voltage with a Test Panel in MAX.  If you still see 5V, it's very likely that the device is damaged and needs to be sent in for repair - check out this KB for instructions.
Cody A.
Message 4 of 8
(4,550 Views)
How can I do the shorting AI0 to GND? Thanks.
Message 5 of 8
(4,503 Views)
You will need to physically wire them together. Can I ask what terminal block you are using with the 6110?
Dale S.
Message 6 of 8
(4,494 Views)
I am using the BNC 2110 block.
Message 7 of 8
(4,491 Views)
The simplest way would be to use a Grounded reference source configuration and wire the center conductor of the AI 0 port to its outer shield.
Dale S.
Message 8 of 8
(4,488 Views)